ALCON - ORLANDO TECHNOLOGY CENTER LADARVISION 4000 OPHTHALMIC EXCIMER LASER SYSTEM

Catalog Number 8065490084
Device Problem Tracking
Event Date 01/26/2006
Event Type Malfunction
Event Description
The tracker stopped tracking a patient's eye at 88% during a customer lasik procedure. They were unable to complete the procedure within the same session, however, there was no patient impact/injury associated with this event.

Manufacturer Narrative
H. 3. , h6. Evaluation summary: the field service engineer was dispatched to the site to evaluate the device. He was able to duplicate the difficulty with the tracker not tracking. He replaced the galvo box, dsp card, analog box, and completed an alignment of the tracker. Then he completed a system verification. The replaced parts were received in manufacturing, but the evaluation has not been completed.

http://www.accessdata.fda.gov/scripts/c ... _ID=680423


Top
 
Display posts from previous:  Sort by  
Post new topic Reply to topic  [ 1 post ] 

All times are UTC


Who is online

Users browsing this forum: No registered users and 1 guest


You cannot post new topics in this forum
You cannot reply to topics in this forum
You cannot edit your posts in this forum
You cannot delete your posts in this forum
You cannot post attachments in this forum

Search for:
Jump to:  
cron
Powered by phpBB © 2000, 2002, 2005, 2007 phpBB Group  
Design By Poker Bandits