 Post subject: FDA & US Marshalls Seize Illegal Laser from Dr Nick Caro
PostPosted: Fri Mar 31, 2006 4:29 am 
Offline
 Profile

Joined: Sat Dec 31, 2005 3:01 pm
Posts: 96
DR. NICHOLAS CHARLES CARO & SAINT GEORGE CORRECTIVE VISION CENTER, SEIZED LASER WITH FULL U.S. MARSHALL SQUAD...
U.S. Marshalls Seize "Homegrown" Laser Owned by Dr. Nicholas Caro
Food and Drug Administration
2098 Gaither Road
Rockville, MD 20850

WARNING LETTER

Nicholas C. Caro, M.D.
4145 W. Peterson, Suite 200
Chicago, Illinois 60640

Dear Dr. Caro,

The Food and Drug Administration (FDA) inspected your medical practice
located at the above address on April 23, 1997, and determined that you
have an unapproved excimer laser system, which was manufactured by
Photon Data, Inc., Winter Park, Florida and assembled by [name redacted
by FDA] Excimer Laser systems are considered to be devices within the
meaning of section 201(h) of the Food, Drug and Cosmetic Act (the Act).
Excimer laser systems are Class III devices which are required to have
in effect an approved application for premarket approval (PMA) or an
approved Investigational Device Exemption (IDE).

On January 24, 1997, you submitted an application for an IDE for your
excimer laser system for use in refractive eye surgery. On February
14, 1997, the Office of Device Evaluation (ODE) at FDA's Center for
Devices and Radiological Health (CDRH) sent you a letter disapproving
your IDE application. You may not use your excimer laser system to
treat human subjects until you have received either an approved PMA
under section 515(a) of the Act, or an approved IDE under section
520(g).

Your excimer laser system is adulterated under section 501(f)(1)(B) of
the Act because it is a Class III device under section 513(f), which is
required to have in effect an approved PMA or an approved IDE, and no
such PMA or IDE is in effect for it. Medical devices used by
physicians in the course of their practice to treat patients are
"marketed" and "held for sale" within the meaning of the Act. Further,
your continued use of this device to treat patients is also a violation
of the Act.

In addition, your excimer laser system must comply with the
requirements of Federal Performance Standards for lasers which are
found in 21 CFR Title 21 of the Code of Federal Regulations, Part 1040.
We acknowledge receipt of a Laser Product Report from you for this
excimer laser system. However, FDA found this Report to be deficient.
Therefore, your excimer laser system is in violation of the Federal
Performance Standards for lasers.

Please note that FDA does not consider your excimer laser to be a
custom device. Section 520(b) of the Act establishes five conditions,
each of which must be met by a device to be a custom device. The Act's
custom device definition requires that the device be made to meet
either the specific anatomical requirements of an individual patient or
the special needs of an individual practitioner; a practitioner's
special needs may be either an individual anatomical need or a special
practice need that is not shared by other physicians.

We do not believe the requirements of your medical practice are unique
because they are shared by numerous other health professionals. In
addition, we do not believe your device is designed to meet any special
anatomical needs that you or an individual patient of yours may have.
Accordingly, your laser is not a custom device and is not exempt from
the requirement under the Act that this device must have an approved
PMA or IDE in effect.

Please notify this office within 15 working days of your receipt of
this letter as to what, if any, actions you are taking or plan to take
to bring your device into compliance with the Act. Your response
should also clearly state whether or not you have ceased using the
device to treat patients. Failure to immediately and completely cease
clinical use of the device upon receipt of this letter, and failure to
bring your device into compliance with the Act, may result in
regulatory action by FDA without further notice. These actions
include, but are not limited to, seizure, injunction, and/or civil
penalties. Please note that no extensions of the 15 day response
period will be given.

Your response should be sent to the attention of Mary-Lou Davis,
Dental, ENT and Ophthalmic Devices Branch (HFZ-331) at the letterhead
address. In addition, please send a copy of your response to Mr.
Stephen Dr. Eich, Compliance Officer, Food and Drug Administration, 300
S. Riverside Plaza, 5th Floor, Suite 550 South, Chicago, Illinois
60606. If you have further questions, please contact Mary-Lou Davis at
(301) 594-4613 extension 127 or FAX: (301) 594-4638.

Sincerely yours,
Lillian J. Gill
Director
Office of Compliance
Center for Devices and Radiological Health

Les Weinstein
Ombudsman
Office of the Center Director
Center for Devices and Radiological Health
Food and Drug Administration
301-827-7991
Fax: 301-827-2565
les.weinstein@fda.hhs.gov (new address)


Top
 
Display posts from previous:  Sort by  
Post new topic Reply to topic  [ 1 post ] 

All times are UTC


Who is online

Users browsing this forum: No registered users and 1 guest


You cannot post new topics in this forum
You cannot reply to topics in this forum
You cannot edit your posts in this forum
You cannot delete your posts in this forum
You cannot post attachments in this forum

Search for:
Jump to:  
cron
Powered by phpBB © 2000, 2002, 2005, 2007 phpBB Group  
Design By Poker Bandits